Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.
Zakir Hussain
Toshiba Sonolayer SSH-140A Error
June 22, 2007 04:29
We are a charity hospital in Skardu. We have a Toshiba Sonolayer SSH-140A for sometime now. Recently we are unable to use it. It displays the following error:
TR ERROR please power off
How we can resolve this issue? Another question is, where we can get manual of this machine? We will be thankful for any help.
to rate and post a comment
Lakshmi Narayana
Re: Toshiba Sonolayer SSH-140A Error
June 22, 2007 02:37
Are you getting this T/R arror with all the probes? Normally T/R error comes with faulty probe. Or pulser board may be bad. Check analog boards.
to rate and post a comment
kenan celik
Re: Toshiba Sonolayer SSH-140A Error
June 27, 2007 02:32
your problem is ;
1 transmitte receiver boards error. T/R board.
2 power supply higt voltage error (dc 90 volt)
3 pulser boards error
4 prob selector boards error
you must chek all
to rate and post a comment
Augustine Kobla Amenyo
Re: Toshiba Sonolayer SSH-140A Error
June 09, 2011 10:13
How to change the language from German to English?
Please help!!
E-mail address: ecesamenyo@yahoo.com
to rate and post a comment
Vu Toan
Re: Toshiba Sonolayer SSH-140A Error
March 18, 2008 02:53
I thinks your unit has problem in RECEIVE Board
You can adjs SWs on the RECEIVE board can cleard this ERROR
to rate and post a comment