Dear All
Spiral Toshiba CT shows the following error during the high kv scan (135 kv and ubove)
Error in XC
Code[0x62640000]
Dev [0x00010002] pno[1]
Any body explain the a.m. error.
yes I am, most high KV XC errors are caused due to a arc my first step is verify the HV cables the HV grease on these systems have a issue with drying out and breaking down if these ends are good then I would recommend running the XC maintenance program to figure out if it is the transformer, cable or tube.
Lightly used system available direct from the end user customer. MONARCH QUEST
Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
Borhan Kalash
Toshiba XC error
February 14, 2014 05:36
Dear All
Spiral Toshiba CT shows the following error during the high kv scan (135 kv and ubove)
Error in XC
Code[0x62640000]
Dev [0x00010002] pno[1]
Any body explain the a.m. error.
to rate and post a comment
Lenny Place
re: Toshiba XC error
February 15, 2014 10:23
have you repacked the HV coupling gel
to rate and post a comment
Borhan Kalash
re: re: Toshiba XC error
February 16, 2014 05:19
Dear Lenny
thanks
Do you mean HV isolater grease?
Do this error means ARC?
thanks in advance
to rate and post a comment
Lenny Place
re: re: re: Toshiba XC error
February 17, 2014 10:05
yes I am, most high KV XC errors are caused due to a arc my first step is verify the HV cables the HV grease on these systems have a issue with drying out and breaking down if these ends are good then I would recommend running the XC maintenance program to figure out if it is the transformer, cable or tube.
to rate and post a comment