if it have some times x-ray & some times no , check hv trans -hv cable & tube. if you hear strange voice(arc voice) , make dark room & exp, You can see the arc place. if not seen, disassemble collimator & set tube window toward control room & do exp.if see any arc in x-ray lamp, it must be change.
best regards
Hello, I need schematic diagram of the equipment SIEMENS Gigantos or Tridoros 5S, need only regard to the autotransformer and generator. My email is robsonquinaglia@gmail.com Best regards.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
sara ahmed
Tridoros-5s
April 28, 2016 11:08
Dear all
I have problem in siemens tridoros -5s machine .so i need service manual.
Pls, any help.
to rate and post a comment
Lenny Place
re: Tridoros-5s
April 29, 2016 01:17
i might have a manual that would help you need an email address
to rate and post a comment
sara ahmed
re: re: Tridoros-5s
May 01, 2016 06:39
sara.ahmed.hazem@gmail.com
Best regards
to rate and post a comment
MIGUEL GONZALEZ
re: Tridoros-5s
April 29, 2016 01:24
very old equipment, what problem you have?
i have manuals of siemens gigantos e, is the same
to rate and post a comment
Juancho
re: Tridoros-5s
April 29, 2016 04:44
Hi, Sara
What is your problem? Please send an email, I have it and I can help you to fix the problem.
Thanks
to rate and post a comment
sara ahmed
re: re: Tridoros-5s
May 01, 2016 06:39
Dear friend
I apprecite your reply.
Really i have problems in two machines of siemens tridoros 5s.
First one:
When i push ex-button anode is rotating and there is not x-ray.
I check hand switch ,it is ok.
Second :
When i push ex-button i hear strange voice and sometimes there is not x-ray.
I tkink it needs to recalibrate the main tranformer voltage.
Thanks for every thing.
Sara.ahmed.hazem@gmail.com
to rate and post a comment
morteza sharifi
re: re: re: Tridoros-5s
May 02, 2016 09:17
if it have some times x-ray & some times no , check hv trans -hv cable & tube. if you hear strange voice(arc voice) , make dark room & exp, You can see the arc place. if not seen, disassemble collimator & set tube window toward control room & do exp.if see any arc in x-ray lamp, it must be change.
best regards
to rate and post a comment
morteza sharifi
re: Tridoros-5s
May 01, 2016 06:29
hi sara ahmad.
how are you & your family?
I have some wiring diagram of trdoros 5 & gigantos , if is good I send it . these are same
to rate and post a comment
sara ahmed
re: re: Tridoros-5s
May 02, 2016 09:18
Dear Morteza sharifi
Iam and my family fine.thanks
You can send it ,pls.
God bless you .
sara.ahmed.hazem@gmail.com
to rate and post a comment
amed yacoub
re: re: Tridoros-5s
November 08, 2016 10:23
Hello morteza sharifi can you send me this wiring diagram of tridoros & gigantos my email is yacoubahmed94@gmail.com Best regards
to rate and post a comment
Robson Quinaglia
re: re: re: Tridoros-5s
November 09, 2016 10:21
Hello, I need schematic diagram of the equipment SIEMENS Gigantos or Tridoros 5S, need only regard to the autotransformer and generator. My email is robsonquinaglia@gmail.com Best regards.
to rate and post a comment