Trouble DARC Computer GE light speed 4slice

January 22, 2016 10:23

Hello
Some help me with CT GE light speed, the system no working because every time turn on the monitor show me error.
"ping darc: waiting....
and then
" failure - waiting for ping darc response (exit status:8)
I conected monitor darc computer I can see the system up but after a few minutes show me...
"see you on the darc side of the moon .
darc login:
...."
thanks

1 person liked this
Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: Trouble DARC Computer GE light speed 4slice

January 23, 2016 10:35

Darc is working if you see that message.

Log inor Register

to rate and post a comment

(66)
(5)

Mr.CT.service

re: re: Trouble DARC Computer GE light speed 4slice

February 19, 2016 08:45

Hi,

I having problem with Darc of LS16. Host is a XW8200.
Sometimes It does not power on with "service darcpower start" command. With Darc manual on no ping neither between them. No ping from darc to hosk neither from host to dark never. Pinging from my LAPtop Host and Darc both are ok.
After several connect/discorect network cable sytsem is fine. When ping is ok every console command to darc is ok.

When power on darc? always on or power on by Host? How is setting Darc bios? just WOL and/or always power on too.

Everything works ok for few days then again unable to power on Darc from host with "service darcpower start" commnad, no pinging between them. Everything works after several connect/disconnect network cable.
This short cable that connects Host and darc, does it have to be a crossover cable? or both Network cards are auto-MDIX?

BTW which darc port is network cable connected to? Darc has two OnBoard RJ45 connector.

Any suggestion is appreciated.

Best regards.

Log inor Register

to rate and post a comment

Gilberto Rubio Gil

re: Trouble DARC Computer GE light speed 4slice

January 23, 2016 10:35

Your Darc is booting normaly, check ethernet connection from hp to Darc

Log inor Register

to rate and post a comment

(148)
(25)

Germni

re: re: Trouble DARC Computer GE light speed 4slice

January 25, 2016 01:34

ok,
I check the ethernet with ping and DARC not have comunication, I reload the IP and the system is working.
thanks

Log inor Register

to rate and post a comment

(66)
(5)

Mr.CT.service

re: re: re: Trouble DARC Computer GE light speed 4slice

February 19, 2016 08:45

Hi,
I'm new with GE CT scanner.
What do you mean with reload IP?
IP at darc or Host?
How do you do it?

Best regards

Log inor Register

to rate and post a comment

(70)
(11)

Bruce Barton

re: Trouble DARC Computer GE light speed 4slice

January 23, 2016 10:38

Network cables from the main system to the darc? Is the darc rebuilding the raid when booting and slowing down the boot process? Can the main computer ping the darc?

Log inor Register

to rate and post a comment

Eric Hauk

re: Trouble DARC Computer GE light speed 4slice

January 23, 2016 10:38

Have you actually tried to manually ping the Darc? If you cannot manually ping the darc you likely have a bad NIC card either on your HP or darc computer, check for link lights on both

Log inor Register

to rate and post a comment

CRAIG WAMBURA

re: Trouble DARC Computer GE light speed 4slice

January 25, 2016 09:58

Try and reload the Serial Over LAN (SOL) CD

Log inor Register

to rate and post a comment

(17)

Gustavo Sierra

re: Trouble DARC Computer GE light speed 4slice

October 05, 2017 09:05

Hi German, I have xpereience en this CT. I can you help. my email. gustavosierra2007@gmail.com

Log inor Register

to rate and post a comment

(17)

Gustavo Sierra

re: Trouble DARC Computer GE light speed 4slice

October 05, 2017 09:05

Hi German. I need de software and aplication disk 1 (2395160) aplication disk 2 (2395161) Patch disk (2399867) and IRIX. Set software H2M5. Can you Help me? thanks.
my email. gustavosierra2007@gmail.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Brand new OEM Philips & Siemens Parts at unbeatable prices

Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty! Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)