April 27, 2008 07:07

hi
i have a problem with an ALT ultramark 9 the sytem goes through boot sequence, after 4 minutes , it displays a message "system controller not responding" pls if anyone have the solution to that probleme help me

Log inor Register

to rate and post a comment

Dave Albaugh

Re: Ultramark 9 " system controller not responding"

April 28, 2008 09:54

Extend the control panel out, behind it you will find a brown plug with three wires, disconnect this plug, this removes power from the control panel. Try to reboot the system. If the system boots the problem is in the control panel

Log inor Register

to rate and post a comment

(158)
(30)
(1)

tahar nouri

Re: Ultramark 9 " system controller not responding"

April 28, 2008 02:31

firt of all , thank for help
the problem was in floppy disk system1
but i got a new problem , ther si no beam on
screen, i changed the main beam former card
MNBF ,but allways no beam ; pls help me.

Log inor Register

to rate and post a comment

(6)
(1)

tony bullon

Re: Ultramark 9 " system controller not responding"

November 17, 2008 04:07

hello,
i have the exact same problem. i disconnected the panel and the system still does not boot up. somebody told me that it could be the floopy disk drive but there are 2 floppy drives on this machine. could this be the source of the problem?
thanks.

Log inor Register

to rate and post a comment

(67)
(3)
(1)

Lobsang Ortiz

Re: Ultramark 9 " system controller not responding"

July 19, 2011 12:53

I have the same problem, i think is the floppy drive and the software disk, is anybody has the software disk?? lobsang_ortiz@hotmail.com
Regards

Log inor Register

to rate and post a comment

(5)
(2)

ali koƧ

re: Re: Ultramark 9 system controller not responding

January 09, 2013 09:12

HELLO THERE BENDE Ultramark 9 ultrasound system. PROBE FAILURE OF VIEW THERE IS A RIDICULOUS unassisted, could you PLEASE e-mail: akdereli17@gmail.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: C2 Management Surplus Auctions July 21st

C2 Management closeout auctions take place on DOTmed.com. Bid today on 345 lots of medical equipment. Onsite or virtual inspection available. Appointments may be made through email at auctions@tryc2.com, or through phone at 540-955-5740.

stats
advertisement

Upcoming Auction Events

PBC Medical Solutions
TIMED Ending Soon
Jul 17, 2026 (11 lots)
C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)