Danial Rao
Universal CP700 x-ray Generator
February 10, 2011 04:53
we are installing a CP700 x-ray generator but after the connections,display is ON and varies the MAS or KV values but when we press the prep button,Filament become glow in tube,rotor runs but it will not become ready for Exposure,any help regarding this will be very appreciated.
to rate and post a comment
Jose Vallejos
Re: Universal CP700 x-ray Generator
February 12, 2011 08:39
You first calibrate with genware software?
You may not have set well the type of x-ray tube or the type of stator, therefore, the equipment does not correctly senses the proper speed of rotation, so i do not give way to X-ray shot.
Good Luck.
Jose
to rate and post a comment
Richard Orlando Suarez Hidalgo
re: Re: Universal CP700 x-ray Generator
October 05, 2016 05:00
Good afternoon. I have a problem with cp 700 computer that generates a short turn me on the dance card charge current regulator between point 3 and point 4 t2 t1. I 120 v appear between these points do not understand because only bone there must be a return 0. please appreciate any help you
to rate and post a comment
Mark Graham
re: Universal CP700 x-ray Generator
June 11, 2012 09:13
I'm having a similar problem but with a CP700 that has been working and then stopped. Everything I check checks out good and the manual I have does not explain how the rotor sensor circuit works. Changed boards, cables, tube, still no ready. How did you solve your problem.
to rate and post a comment
Jose Garcia
re: re: Universal CP700 x-ray Generator
July 10, 2017 09:14
Hi Can I get a copy of this service manual?
to rate and post a comment
Jose Garcia
re: Universal CP700 x-ray Generator
July 10, 2017 09:14
Hi Do you have service manual of this equipment?
to rate and post a comment
Jose Garcia
re: Universal CP700 x-ray Generator
July 10, 2017 09:16
Do you have the service manual ?