I have a Universal MP500 - 30kW generator. The maximum kV is limited to 99 kV. I want to configure this to 125kV. Does anybody know how to do this?
I have service manual 8000-MP500 rev A (feb 9 1995)- but no discription how to configure kV max.
Thanks for your reply.
The kV selection on the console is limited to 99 kV at the moment. My customer want to dail up to 125 kV maximum in future. Is there a possibility for a hardware configuration
There is an Eprom to de-rate the generator, but that normally does not limit the kV. It would limit the mA and the Eproms would de-rate from 30kW to maybe 18kW. It sounds like a different problem, maybe something on the main board in the operator console.
I can send you the manual. However, nothing states in the manual about changing tube type. It says to consult the factory for this. I do not know how it would be changed, but I suspect it probably requires a change of an EEPROM, probably to drive the filaments. The list of tubes compatiable is over 30 tubes in the manual. The generator really doesn't know what type of tube you may be using. It has no type of settings or selections for different tubes. Again, just that are able to adjust the mA waveform to flat, across the top. I can still send you the manual if you need it. Galen
I would say either a problem with the display/control p.c.b in the operator console or possibly a problem with memory/program p.c.b in the card cage in the generator cabinet.
Thank you for the support.
In fact there is no error in the generator, the system is working properly. 'Somebody' just limited the generator to 99kV (for radiation law reason). Now the customer asked for higher kV settings. In the limited service documentation I have there is no indication that the kV can be limited to
I have done service of this model generator since they were first manufactured. The generator does not have any field service configuration or modification mode that would you set or change the maximum kV limit. At least not that I am aware of. There are changes that could be made to configuration through a series of dip switch settings, but they are settings for disable/enable AEC, table bucky, wall bucky, etc. Also there are power on defaults as to what the displays will show at power on. For instance, you can set the dip switches so that the displays will show 300 mAs and 125 kVp when you power on the generator, however, in your situation I don't believe this would help. It appears that someone had a factory mod of the software done to limit the kV. The software is burned onto an Eprom.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Universal/ DEL MP500 generator
September 26, 2012 10:30
I have a Universal MP500 - 30kW generator. The maximum kV is limited to 99 kV. I want to configure this to 125kV. Does anybody know how to do this?
I have service manual 8000-MP500 rev A (feb 9 1995)- but no discription how to configure kV max.
thanks in advance
to rate and post a comment
Alfred M. Brown
re: Universal/ DEL MP500 generator
September 27, 2012 05:33
Do you mean you can only dial up 99kv on the console or the console has 125kv on it and the actual kv measured is only 99kv?
to rate and post a comment
Rob Spanjaart
re: re: Universal/ DEL MP500 generator
September 28, 2012 01:58
Hi Alfred,
Thanks for your reply.
The kV selection on the console is limited to 99 kV at the moment. My customer want to dail up to 125 kV maximum in future. Is there a possibility for a hardware configuration
to rate and post a comment
Galen Hiveley
re: Universal/ DEL MP500 generator
October 01, 2012 07:34
There is an Eprom to de-rate the generator, but that normally does not limit the kV. It would limit the mA and the Eproms would de-rate from 30kW to maybe 18kW. It sounds like a different problem, maybe something on the main board in the operator console.
to rate and post a comment
EDILBERTO JUAREZ
re: re: Universal/ DEL MP500 generator
March 10, 2017 10:07
Galen may I get the service manual of DEL MP500,
I need to update tube type on generator
to rate and post a comment
Galen Hiveley
re: re: re: Universal/ DEL MP500 generator
March 10, 2017 04:47
I can send you the manual. However, nothing states in the manual about changing tube type. It says to consult the factory for this. I do not know how it would be changed, but I suspect it probably requires a change of an EEPROM, probably to drive the filaments. The list of tubes compatiable is over 30 tubes in the manual. The generator really doesn't know what type of tube you may be using. It has no type of settings or selections for different tubes. Again, just that are able to adjust the mA waveform to flat, across the top. I can still send you the manual if you need it. Galen
to rate and post a comment
Galen Hiveley
re: Universal/ DEL MP500 generator
October 01, 2012 08:24
I would say either a problem with the display/control p.c.b in the operator console or possibly a problem with memory/program p.c.b in the card cage in the generator cabinet.
to rate and post a comment
Rob Spanjaart
re: re: Universal/ DEL MP500 generator
October 03, 2012 02:09
Hi Galen,
Thank you for the support.
In fact there is no error in the generator, the system is working properly. 'Somebody' just limited the generator to 99kV (for radiation law reason). Now the customer asked for higher kV settings. In the limited service documentation I have there is no indication that the kV can be limited to
to rate and post a comment
Galen Hiveley
re: re: re: Universal/ DEL MP500 generator
October 04, 2012 08:45
I have done service of this model generator since they were first manufactured. The generator does not have any field service configuration or modification mode that would you set or change the maximum kV limit. At least not that I am aware of. There are changes that could be made to configuration through a series of dip switch settings, but they are settings for disable/enable AEC, table bucky, wall bucky, etc. Also there are power on defaults as to what the displays will show at power on. For instance, you can set the dip switches so that the displays will show 300 mAs and 125 kVp when you power on the generator, however, in your situation I don't believe this would help. It appears that someone had a factory mod of the software done to limit the kV. The software is burned onto an Eprom.
to rate and post a comment
IngJose
re: re: re: re: Universal/ DEL MP500 generator
July 20, 2016 10:26
Hi Galen this is Manuel from Mexico i am looking for service and operator manual of this x ray system could you help me ?
in advance thanks
to rate and post a comment
Galen Hiveley
re: re: re: re: re: Universal/ DEL MP500 generator
July 21, 2016 10:02
Need your email address. Galen
to rate and post a comment
JUAN MANUEL ROCA SALAZAR
re: Universal/ DEL MP500 generator
June 13, 2016 12:24
Hi, I 'm looking for the manual x-ray equipment UNIVERSAL MP500 HF . Any help please email me atlanta_peru@yahoo.es
thank you very much
Juan Roca
to rate and post a comment
JUAN MANUEL ROCA SALAZAR
re: Universal/ DEL MP500 generator
June 13, 2016 12:25
Hi, I 'm looking for the manual x-ray equipment UNIVERSAL MP500 HF . Any help please email me atlanta_peru@yahoo.es
thank you very much
Juan Roca
to rate and post a comment
Galen Hiveley
re: re: Universal/ DEL MP500 generator
June 13, 2016 06:10
Hello Juan. I sent the MP500 service manual to your email address earlier today. Regards, Galen