Dear collegues,
does anyone know, should ups-battery warning leds stay lighting after changing of the batteries. Particulary in the "Little Star" Toshiba ups. I found, that for example, an APC ups makes battery test once a 14 days and from the test the leds will be set.
Iosif,
The Little star UPS is tricky, if you let the battery drain too far down, it will delete the firmware inside the UPS unit itself. After that, it (usually) will not work properly again.
I suggest replacing the Little Star with an APC 700 or 800.
Worst case, you will have to manually startup the UPS after powerdown (my sites do not power down, so its not a problem.
Little Star is junk after 8-9 years.If you look at the wiring on the back,you can eliminate it all together.Little star is only used if power to building goes off giving you a chance to 'shutdown' the system.
yes, I know it's simple, but unfortunately the system is installed in place with a lot of probability that the power can be often off.
Anyway, thank you.
Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com
Iosif Sogolov
ups mri
January 29, 2018 03:22
Dear collegues,
does anyone know, should ups-battery warning leds stay lighting after changing of the batteries. Particulary in the "Little Star" Toshiba ups. I found, that for example, an APC ups makes battery test once a 14 days and from the test the leds will be set.
Thank you in advance.
to rate and post a comment
Nicholas Cipriano
re: ups mri
February 23, 2018 04:06
Iosif,
The Little star UPS is tricky, if you let the battery drain too far down, it will delete the firmware inside the UPS unit itself. After that, it (usually) will not work properly again.
I suggest replacing the Little Star with an APC 700 or 800.
Worst case, you will have to manually startup the UPS after powerdown (my sites do not power down, so its not a problem.
to rate and post a comment
Iosif Sogolov
re: re: ups mri
February 26, 2018 09:52
Thank you Nicholas
to rate and post a comment
alex malk
re: re: ups mri
March 07, 2018 09:08
you still working for toshiba/cannon.? i know you from a Vantage class i think.
to rate and post a comment
alex malk
re: ups mri
March 07, 2018 09:08
Little Star is junk after 8-9 years.If you look at the wiring on the back,you can eliminate it all together.Little star is only used if power to building goes off giving you a chance to 'shutdown' the system.
to rate and post a comment
Iosif Sogolov
re: re: ups mri
March 07, 2018 04:50
yes, I know it's simple, but unfortunately the system is installed in place with a lot of probability that the power can be often off.
Anyway, thank you.