Over 500 Total Lots Up For Auction at One Location - NJ 06/15

US vivid3

(56)
(15)

imeng

US vivid3

September 25, 2014 10:23

Hi everybody,
We have vivid3 that doesnt bootup. the onoff button on kB is lit. after pressing On, the system funs and lEDs on BEP and FE are lit, DVD lit, but the screen stay balck. I have tested the Screen with my laptop the VGA display ok. I tried to boot from CD sw but nothing on screen, i have tested the HDD is ok.
any Idea? Thank you!
regards,
Imeng

Log inor Register

to rate and post a comment

(32)
(4)

Tiorico_86

re: US vivid3

September 26, 2014 10:11

You must to check the Front End DC Power Supply (LVPS), check for a VDR burnt, dissconnect the cable PC2IP from PC2IP board and check if the monitor doesn't remains black(with video signal on it)the fault is the LVPS, if remaind black check on the BEP.
Good Luck.

Log inor Register

to rate and post a comment

CHRIS NALMPANTIS

re: US vivid3

September 26, 2014 10:14

this sounds like motherboard from pc. try to change it.It is a INTEL D845EPI/D845GVSR