Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Videomed Sx

(20)
(2)

Ogali Daniel

Videomed Sx

September 07, 2012 06:49

I am about to complete installation of Axiom iconos R-200. When fluoroscopy is triggered,no image appears on TV screen. I found message from D10 saying, camera tube filament power supply out of tolerance. Also +6.5v LED on D1 was dimmer than other LED. The filament voltage was +3.8v instead of +6.5v. I changed the D1 board, the filament power supply was ok,image ok on TV,but after some time the same problem came up with D1 board. Where is the cause of this? I have cried out ohmic checks from D12 x200 camera cable connector, ok. Also voltage checks on D1 board test points, X204,x205,etc. Can D10 be faulty? Please help.

Log inor Register

to rate and post a comment

sohail aftab

re: Videomed Sx

September 08, 2012 09:32

please send me your email add or contact supermeditech@yahoo.com

Log inor Register

to rate and post a comment

(20)
(2)

Ogali Daniel

re: re: Videomed Sx

September 08, 2012 04:29

my e-mail add: codemexee@yahoo.co.uk