VIVID 7 DOPPLER/IMAGE
January 20, 2015 10:16
Hello, I have the following problem on a VIVID 7: - when I turn on the doppler test, on the screen appears a green scramble, something like a noise interference. I checked the power supply, thinking it might be a problem of power fluctuation, so I plugged it to an UPS, that didn't solve the problem. Can anyone sugest something that I might try, or if someone had this problem? Thx!
to rate and post a comment
Rey Gonzalez
re: VIVID 7 DOPPLER/IMAGE
January 20, 2015 06:05
run system diagnostic. rey@allultrasoundservices.com
Marcelo Antonio Correa
January 21, 2015 09:49
Hi, Horia! Last time I faced such artifact on CWD (Nov, 2009), I replaced the board RX128-4, p/n FC200057-06. Hope this helps, take care, God Bless. Kind regards, MAC.
Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com
horia mititelu
VIVID 7 DOPPLER/IMAGE
January 20, 2015 10:16
Hello, I have the following problem on a VIVID 7:
- when I turn on the doppler test, on the screen appears a green scramble, something like a noise interference.
I checked the power supply, thinking it might be a problem of power fluctuation, so I plugged it to an UPS, that didn't solve the problem.
Can anyone sugest something that I might try, or if someone had this problem?
Thx!
to rate and post a comment
Rey Gonzalez
re: VIVID 7 DOPPLER/IMAGE
January 20, 2015 06:05
run system diagnostic. rey@allultrasoundservices.com
to rate and post a comment
Marcelo Antonio Correa
re: VIVID 7 DOPPLER/IMAGE
January 21, 2015 09:49
Hi, Horia!
Last time I faced such artifact on CWD (Nov, 2009), I replaced the board RX128-4, p/n FC200057-06.
Hope this helps, take care, God Bless.
Kind regards,
MAC.