Leonid Seidlitz
Vivid 7 pro
July 30, 2021 10:33
Hello, on vivid 7 pro two linear probes with more than 128 transducer elements shows no echo or Doppler. Another probes are good in any mode. These linear probes work good on on another vivid 7. No errors during tests on RFI, TXP, TX. Maybe it is TXP, RFI or RLY Boards fault?
Any suggestions are appreciated.
Regards,
Leonid
to rate and post a comment
JohnDang
re: Vivid 7 pro
August 18, 2021 03:27
Hello,
are there same model of Linear probes ? becasue if they are different elements, maybe your board is gotten problem. Please try to reload sw first if you have.