vivid i

emad aly

vivid i

January 28, 2014 12:04

hello my friends
i have a problem in vivid i machine
the machine was working fine and it had left for 4 days. today i turned it on it works but says no probe and there are 5 blanking circles in the us image
the probes are fine (3S and 8L)
i need help
emadmekawy@gmail.com

Log inor Register

to rate and post a comment

Marcelo Antonio Correa

re: vivid i

January 29, 2014 08:19

Hi, Emad! I looks like bad contact to me.

In bigger equipments you see this kind of behaviour when FEC board losts communication to BEP and, many times, it's just bad contact.

I'd open the equipment and would reattach all boards.

Depending on the equipment's age and climate conditions, such as temperature and humidity, they are very responsible for such failure, sometimes.

Hope this helps. Take care, God Bless.

MAC.

2 people liked this
Log inor Register

to rate and post a comment

(165)
(67)
(2)

Sergey Zheleznyak

re: vivid i

January 29, 2014 08:31

Hello, Emad
Looks like your machine is booting into Frontend simulation mode. Usually it happens then you have internal communication problems.
If you can collect logs from your system - that might be helpful. Try Alt+D combination and export to external USB stick.
Also sometimes wrong BIOS settings prevent system from booting correctly. Press F2 "Config/Diag" button after system power-on. If system lets you enter BIOS without password - you need to change BIOS battery and configurate BIOS properly. If system asks for the pass - that's not your case most likely - start collecting and analyzing logs.
Good luck.

Log inor Register

to rate and post a comment

emad aly

re: vivid i

January 29, 2014 03:27

Hello my friend
Thank you for supporting me
after cleaning sockets and use hot air the machine works fine
Many thanks

Log inor Register

to rate and post a comment

Robert Harris

re: re: vivid i

January 30, 2014 11:44

Yes it was probably condensation. Also be real careful with leaving these units in hot like in cars. I have seen the screens bubble and drip liquid out of them