Voluson 530D MT Imaging Error scrambled image

Tags: Ultrasound, MEDISON

(7)
(3)

Medimarc

Voluson 530D MT Imaging Error scrambled image

August 09, 2015 09:29

Dear readers,

I have a problem with a Voluson 530D MT machine that shows a scrambled picture on the monitor - I already checked in the service menu via PC keyboard that it cannot be in the PC unit below, what could be the reason?

Log inor Register

to rate and post a comment

(8)
(2)

jason_ampronix

re: Voluson 530D MT Imaging Error scrambled image

August 11, 2015 04:22

Can you try using a secondary monitor and see if the image is still scrambled from your Voluson 530D MT? If it is scrambled, that is coming from your system. If the image is not scrambled, that means the original monitor has issues.

Please let me know how this goes.