ROBERTO DIAZ BEDOLLA
voluson 730 does not start
September 20, 2011 02:15
HI, I HAVE A PROBLEM WITH VOLUSON 730,..WHEN U TURN IT ON THE TOUCH SCREEN ONLY SHOWS "voluson 730" ALL FANS WORK AND THE MAIN SCREN IT SEEMS IT'S GOING TO START BUT THAT'S ALL... NEVER HAPPENS ANYTHING...
YESTERDAY I CHECK THE VOLTAGE IN THE HARD DISK AND AS U KNOW I HAVE 12V BUT INSTEAD OF 5V (ON THE OTHER CONECCTION) I HAVE JUST 1.4
I HOPE SOMEONE CAN HELP ME THX
SOY DE MEXICO
to rate and post a comment
Rodolfo Figueroa
Re: voluson 730 does not start
September 21, 2011 09:04
todo parece que tu fuente se quemo es el gabinete de abajo donde se connecta a la luz.
you got a bad power supply is the bottom module where is located the power cord,
to rate and post a comment
THANH TRUONG
Re: voluson 730 does not start
September 21, 2011 09:04
At first, check all DC voltages on power supply,
If it is good, check solfware in harddrive, may be solfware or harddrive failwd
to rate and post a comment
shawn stan
Re: voluson 730 does not start
January 27, 2012 04:52
I need help please. I just bought a voluson 730 from 2002. It worked fine for about 2 weeks. Now when i boot up the system the voluson logo appears on the touch screen, the monitor powers up but no image. nothing happens after this, it just stays like that. when i turn off the machine, it wont turn off unless i trip the fuse or unplug the machine from the wall. when this all statred to happen i thought i heard a faint pop coming from the machine. can anybody help me??
I also dont see any cd's or dvd's that came with the machine.
to rate and post a comment
STEFANO CIATTI
re: Re: voluson 730 does not start
March 14, 2013 09:27
Shawn , hello
this morning I have the same problem of yours, how did u resolve it?
to rate and post a comment
Seba Bek
re: re: Re: voluson 730 does not start
May 08, 2014 04:58
i have the same problem did u solve it ?
to rate and post a comment
Victor Lupusor
Re: voluson 730 does not start
September 21, 2011 09:05
Hi,
open the back cover on the Voluson, you will find 6 green led's (look down on the right). Power on the unit and see if they are light.Configuration is:
-(+15V)--(+3,3V)
-(-15V)--(+5V)
-(FAN) --(+12V)
Inside of the board where the led's are located you will find 2 fuses one for 12V and one for 5V
Hope it helps
Best regards,
Victor
to rate and post a comment
ROBERTO DIAZ BEDOLLA
Re: voluson 730 does not start
September 21, 2011 04:31
thanx.. i already checked it the 5V supply it seems it's the grown
to rate and post a comment
Gilberto Rubio Gil
Re: voluson 730 does not start
September 22, 2011 12:03
If the fuses are OK.
Then your problem is the power supply board.
It is located next to the hard drive.
If you need help I am in Guadalajara
to rate and post a comment
morteza yousefi
re: voluson 730 does not start
November 01, 2013 05:24
hi.i see your texts
I dond understant you have the 5 volt or no.if have 5 volt your sbc bord (moderbord) is defectiv.or you can check the voltag of pwer supply .it must be 59 v fix.and if all 6 lamp of back of bep is on you must cheng the moderbord or repair the moderbord.
to rate and post a comment
babusgajelli Gajelli
re: voluson 730 does not start
May 09, 2014 01:54
Just u need check the CPP board by applying external DC 40v min
Then check output dc(12,+/_15,+/_5,3.3)V
I hope it's 5volts may be the problem if you already know this ignore it
regards
BABUSGAJELLI
to rate and post a comment