Voluson E6 error

June 28, 2018 01:28

Hi forum members
I have done full backup as part of the annual service. System throwing power management error after boot up. I have removed all the probes and rebooted. No error. I have connected a probe and power management error showing on the screen immediately pressed probe button. Any idea which part of the unit throwing this error...

Log inor Register

to rate and post a comment

(74)
(10)
(1)

Javier Abdulov

re: Voluson E6 error

June 28, 2018 03:03

have poblem on power supply maybe,lso you must reload system

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: Voluson E6 error

June 29, 2018 08:39

check connections on probe connector board and power supply

Log inor Register

to rate and post a comment

(34)
(4)
(3)

Thameen Ansari

re: Voluson E6 error

July 01, 2018 12:48

Thank you for Javier and A. Sanchez. software reloaded. Checked all the cables and connections.checked front end boards and power supply modules... All seems to OK. the exact error screen shot as follows.

Error: Error on Power management
Action restart the system
Error in File.\I_Begrenzer\TxPowerToHW_DAST.cpp:436
Version Voluson E6 14.0.8.1685(Oct 2015,15.00.02)

Also I looked at log file and as follows:

Module:HRS
Symbol Full name:vTxPowerOff
Synbol:vTxPowrOff
Source: ,line 0

I am suspecting the power supply but also concern about front end beam former board..can any one tell me which is near possible after reading the above errors..

I have basic service manual.. not helping much ..please share a copy of proprietory manual if any forum members has one.. thank you in advance for further advice..

Log inor Register

to rate and post a comment

(74)
(10)
(1)

Javier Abdulov

re: re: Voluson E6 error

July 02, 2018 10:24

Hi sir think there is some problem power.supply ics or.controllers.it would.be well.known try inspect power.supply by some electronics repairman.And.there is.one.small possiblity that problem may be sourcing by hard drive.but it is some possiblity.

Log inor Register

to rate and post a comment

(34)
(4)
(3)

Thameen Ansari

re: Voluson E6 error

July 09, 2018 08:44

HI All

Thank you for your tech support... finally it is PS issue..

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)