Voluson E8 Expert touchscreen problem
December 12, 2016 10:24
I have Voluson Expert BT08.
After normal boot the touchscreen remaning blue with GE logo and respective US patents.
does not appear probe- aplication -ect.....
All functions on the keyboard are ok except that
After reboot usually go normaly.
Where is the problem?
to rate and post a comment
ClarityDiagnostics
re: Voluson E8 Expert touchscreen problem
December 12, 2016 04:41
You should try a software reload. If the problem continues, check the USB cable of the touchscreen module.
to rate and post a comment
Anthony Ngo
re: Voluson E8 Expert touchscreen problem
December 13, 2016 10:12
Had similar issue with the E8. Cable connections were all fine, but had to do a SW reload before touchscreen worked again.