Diego Seoane

Voluson E8 Expert touchscreen problem

December 12, 2016 10:24

I have Voluson Expert BT08.

After normal boot the touchscreen remaning blue with GE logo and respective US patents.
does not appear probe- aplication -ect.....
All functions on the keyboard are ok except that

After reboot usually go normaly.

Where is the problem?

Log inor Register

to rate and post a comment

(12)
(6)

ClarityDiagnostics

re: Voluson E8 Expert touchscreen problem

December 12, 2016 04:41

You should try a software reload. If the problem continues, check the USB cable of the touchscreen module.

Log inor Register

to rate and post a comment

(39)
(20)

Anthony Ngo

re: Voluson E8 Expert touchscreen problem

December 13, 2016 10:12

Had similar issue with the E8. Cable connections were all fine, but had to do a SW reload before touchscreen worked again.