Hi all,
This is regarding a Voluson P8 issue. The problem is the machine will work fine with some set of linear and Convex probes. If I used another linear or convex probes, it will not work. Probe switching happens, but scan area is not coming instead we can see the back ground. If I use linear probe in the first port, I can hear relay sound ticking continuously. All of the probes which I used are 100% working. I checked in other machines too. I restored the software, reloaded the software in new hard disk, but no use. Please help. Thank you

Log inor Register

to rate and post a comment

(14)
(1)
(2)

RANEESH PULIYULLAKANDIYIL

re: Voluson P8

October 30, 2016 08:39

Cross check probe connector board

Log inor Register

to rate and post a comment

(165)
(67)
(2)

Sergey Zheleznyak

re: Voluson P8

October 30, 2016 08:40

Hello, Sahal
Are You sure that the probes You tried to use are compatible with Voluson P8 system?
Can You name the models of the probes that aren't working?
Send the errorlog to s.v.zhel@gmail.com
Regards

Log inor Register

to rate and post a comment

(31)
(6)

Sahal Ahammed

re: re: Voluson P8

October 31, 2016 10:52

yes, 100%.I used the same probe only. We are having many probes of the same model and I tried all. Some are working and some are not. This 'not working probes' are working in other machines too.

Log inor Register

to rate and post a comment

(165)
(67)
(2)

Sergey Zheleznyak

re: re: re: Voluson P8

November 01, 2016 10:56

What about software versions of the other machines - are they all the same? Need error log - the answer to Your problem is there.

1 person liked this
Log inor Register

to rate and post a comment

(31)
(6)

Sahal Ahammed

re: re: re: re: Voluson P8

November 02, 2016 10:35

No,software versions are different. how to get error logs? there is no error message popping up automatically. Thank you

1 person liked this
Log inor Register

to rate and post a comment

(165)
(67)
(2)

Sergey Zheleznyak

re: re: re: re: re: Voluson P8

November 03, 2016 10:32

The software versions also might influence support of different transducers. Insert USB memory stick and press ALT+ D to collect logs and system data. Better do the same on the system which recognizes the transducer correctly. So we could crosscheck.

2 people liked this
Log inor Register

to rate and post a comment

(31)
(6)

Sahal Ahammed

re: re: re: re: re: re: Voluson P8

November 04, 2016 01:26

Thanks Sergey. But the thing is I have used the same model. If I used different model, then the software version can play a role. But with the same model transducer, some will work which means software is supporting. Ryt?

Log inor Register

to rate and post a comment

(165)
(67)
(2)

Sergey Zheleznyak

re: re: re: re: re: re: re: Voluson P8

November 07, 2016 10:41

You are welcome, Sahal. If the software is different even on the same model - incompatibility issues are still possible. Such as shorter compatible transducers list on older soft.
Also on Administration/Service/probe eeprom You may see status of all probes connected - either the machine doesn't recognize it or probe id reads out wrong...

Log inor Register

to rate and post a comment

(165)
(67)
(2)

Sergey Zheleznyak

re: re: re: re: re: re: re: Voluson P8

November 07, 2016 10:41

Sorry, Sahal. I didn't understand Your thought at first. You are right if transducers models are same - then soft version isn't important - my advice still the same - read out error logs, get probe eeprom info - all will become clear...

Log inor Register

to rate and post a comment

(31)
(6)

Sahal Ahammed

re: re: re: re: re: re: re: re: Voluson P8

November 08, 2016 10:21

Thanks Sergey. Will see to that.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)