Luke Rzeznik
VOLUSON V730 no reference position motctrl error
April 05, 2016 08:53
Hello.
I have a problem with probe RAB4-8L for VOLUSON V730.
The error i get is "MotCtrl: No Reference Position signal!......"
I took of the body and replaced halotrons (Original honeywell) and tried to fix the position of one magnet/halotron (the bottom one).
Still the same.
The head with crystals go only to half of its "capabilities".
I dont want to disassebly if i dont have to (to fix this problem). I just get off housing for now.
If you have some knowledge about repairing this probes and help me i will be happy to help you with others repairs, softwares, manuals etc.
Hope you give me some clue...
Thankyou
to rate and post a comment
miatip
re: VOLUSON V730 no reference position motctrl error
April 05, 2016 06:09
We can repair for you. my email is muhammed@miatip.com
to rate and post a comment
Luke Rzeznik
re: re: VOLUSON V730 no reference position motctrl error
April 06, 2016 09:14
Thanks but I got somebody in my country. I want to know what is causing this problem...
to rate and post a comment
Stam Skentos
re: VOLUSON V730 no reference position motctrl error
April 06, 2016 02:49
We have seen this problem many times and have repaired several GE 4D probes.You cannot repair it if do not have the experience.Most probably your problem is inside the scanning head . Regards Stam
to rate and post a comment
Luke Rzeznik
re: re: VOLUSON V730 no reference position motctrl error
April 07, 2016 12:26
Ok. That answer bring me closer... Expirience... i want to get it right now. But dont want to necessary damaging probe if the problem is somewere else.
Inside head? Magnet or somenthing else?
Thanks
to rate and post a comment