VOLUSON V730 no reference position motctrl error

Tags: Ultrasound, GE

(10)
(1)

Luke Rzeznik

VOLUSON V730 no reference position motctrl error

April 05, 2016 08:53

Hello.
I have a problem with probe RAB4-8L for VOLUSON V730.
The error i get is "MotCtrl: No Reference Position signal!......"

I took of the body and replaced halotrons (Original honeywell) and tried to fix the position of one magnet/halotron (the bottom one).
Still the same.
The head with crystals go only to half of its "capabilities".

I dont want to disassebly if i dont have to (to fix this problem). I just get off housing for now.

If you have some knowledge about repairing this probes and help me i will be happy to help you with others repairs, softwares, manuals etc.

Hope you give me some clue...

Thankyou

1 person liked this
Log inor Register

to rate and post a comment

(1)

miatip

re: VOLUSON V730 no reference position motctrl error

April 05, 2016 06:09

We can repair for you. my email is muhammed@miatip.com

Log inor Register

to rate and post a comment

(10)
(1)

Luke Rzeznik

re: re: VOLUSON V730 no reference position motctrl error

April 06, 2016 09:14

Thanks but I got somebody in my country. I want to know what is causing this problem...

Log inor Register

to rate and post a comment

(19)
(5)

Stam Skentos

re: VOLUSON V730 no reference position motctrl error

April 06, 2016 02:49

We have seen this problem many times and have repaired several GE 4D probes.You cannot repair it if do not have the experience.Most probably your problem is inside the scanning head . Regards Stam

Log inor Register

to rate and post a comment

(10)
(1)

Luke Rzeznik

re: re: VOLUSON V730 no reference position motctrl error

April 07, 2016 12:26

Ok. That answer bring me closer... Expirience... i want to get it right now. But dont want to necessary damaging probe if the problem is somewere else.
Inside head? Magnet or somenthing else?

Thanks

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We provide sales, service & repair for portable ultrasound systems & probes

If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)