villa

WINSCOPE (Exposure Interlock)

December 10, 2014 08:41

Hello everyone, someone ask for me to check a R/F toshiba room, I dont know the system, they only change the tube from table to ceiling, but not make exposures on table (I notice an indication on control of Exposure Interlock, but the ceiling system make exposures, so I dont know.
If someone has tips, how troubleshoot system? how discard all sensors? If possible service manual
Thank you

Log inor Register

to rate and post a comment

(147)
(22)

Wang Zhixuan

re: WINSCOPE (Exposure Interlock)

December 11, 2014 10:04

You said the WINSCOPE only changed the tube from table to ceiling. Could you change the tube to table again? If the
problem still occure, you should check the table to troubleshoot.