Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

X-Ray generator MPG 80 GE Error code 410

Tags: Radiology - R/F, GE

(2)

Hristo Slavkov

X-Ray generator MPG 80 GE Error code 410

January 26, 2017 10:00

Dear Colleagues,
I have a problem with X-Ray generator MPG 80 GE with Prestige VH device, manufactured 1996. An error code 410 displays on start up.
Can any one tell me what it means and how can I fix it.

Best regards

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: X-Ray generator MPG 80 GE Error code 410

January 28, 2017 03:54

DC_BUS_ERROR DC bus out of range
(dc_bus_fault signal)
classIV
Exposure inhibited
Reset error.
If it reappears, find the
default in running diagnostics
on/off and DC bus
MENU2 TEST F1
MENU2 TEST F3
Then follow the faulty tree
ON/OFF and DCPS

Log inor Register

to rate and post a comment

(48)
(9)

Francesco Didonato

re: re: X-Ray generator MPG 80 GE Error code 410

January 31, 2017 10:24

You are problem on the power Supply not energized contactor for 600 volt

Log inor Register

to rate and post a comment