Dear Colleagues,
I have a problem with X-Ray generator MPG 80 GE with Prestige VH device, manufactured 1996. An error code 410 displays on start up.
Can any one tell me what it means and how can I fix it.
DC_BUS_ERROR DC bus out of range
(dc_bus_fault signal)
classIV
Exposure inhibited
Reset error.
If it reappears, find the
default in running diagnostics
on/off and DC bus
MENU2 TEST F1
MENU2 TEST F3
Then follow the faulty tree
ON/OFF and DCPS
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Hristo Slavkov
X-Ray generator MPG 80 GE Error code 410
January 26, 2017 10:00
Dear Colleagues,
I have a problem with X-Ray generator MPG 80 GE with Prestige VH device, manufactured 1996. An error code 410 displays on start up.
Can any one tell me what it means and how can I fix it.
Best regards
to rate and post a comment
Lenny Place
re: X-Ray generator MPG 80 GE Error code 410
January 28, 2017 03:54
DC_BUS_ERROR DC bus out of range
(dc_bus_fault signal)
classIV
Exposure inhibited
Reset error.
If it reappears, find the
default in running diagnostics
on/off and DC bus
MENU2 TEST F1
MENU2 TEST F3
Then follow the faulty tree
ON/OFF and DCPS
to rate and post a comment
Francesco Didonato
re: re: X-Ray generator MPG 80 GE Error code 410
January 31, 2017 10:24
You are problem on the power Supply not energized contactor for 600 volt