I am looking for parts for Jedi generator. I need a Kv board and rotor board.
After installation new tube for GE Silhouette VR, it has error (error 60 0313) during exposure with large focal spot. Exposure for small focal spot is OK. Have you got any idea what is the cause of that?
In my opinion it is kv board or rotor board.
I think that the problem is fixed in HV Tank or KV control board, the description error in service manual for silhouette, indicated as follows:
Invertar max, ILR current detected
In your case if small filament is ok, the rotor board is ok, I{m sure if you you change the KV control board yuopr system running.
I just checked the Proteus service manual (which is a Jedi generator as well) and agree that the problem is the KV control Bd. and worst case it could be the HV tank.
hello
good work
Contact D x -ray machine v silhouette technical service guide is required.
What could be the problem in our popcorn IGBT device. Mail necexray@hotmail.co I
Thank you in advance
I really don't know GE Silhouette VR, but Jedi used by CT systems also as you know and some cases, Err 60-313h ILR Max current detected may caused by tube. If you have different error with old tube be sure your new tube is ok before order KV board. ( one of the filament may ok, but it will fail after a while if the tube is defective). Please try no load test if that kind of test is available at your system.
Old unit didn't rotare.
Bearing had been broken.
When I conected high voltage cables to the old tube (in order to check errors) Erros was the same but inversly. It mean error is for small focal spot. Large is OK.
Marcin,
This generator has diagnostic software you can us to test things on it. You need the Jedi software and a laptop cabable of running DOS and a serial port and cable. I havent had this problem but I know this unit will be difficult to troubleshoot and calibrate without the software. Good luck!
Deon
dear help me i have a xray system GE XR6000 and there error C03-0001 JEDI Communication Error, the cables is ok, the TAV board is OK, the console is oK too, but the problema follow
please write to aalzamoram@hotmail.com
Thanks
dear help me i have a xray system GE XR6000 and there error C03-0001 JEDI Communication Error, the cables is ok, the TAV board is OK, the console is oK too, but the problema follow
please write to aalzamoram@hotmail.com
Thanks
Hi:
I have a GE XR6000 xray system, and the console shows the C005-0001 error most of the time. I just can take head shots, but when i Change to lungs or chest or spine, this error appears and the anode doesnt even rotate. Sometimes when I change to stand bucky to table bucky allows the shot, but then fails. Can you help me please. I dont have the jedi software nor db9 cable configuration to connect and diagnost well. Please help me with this. My mail is: irina.saenz81@gmail.com
Which software is required? Mostly, there is software to upgrade to Jedi 5.47, but normally, one saves the previous data, then downloads onto non-volatile memory when changing the KV Control.
The manual is incorrect, can send a proper copy to your EMAIL
Hello.
I have an X-ray generator brand Ge 60/80/100 series. The console displays "Ext lock.". Would you help me understand this message? Thank you very much.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Marcin Gorny
X-ray jedi generator problem
February 13, 2007 09:28
I am looking for parts for Jedi generator. I need a Kv board and rotor board.
After installation new tube for GE Silhouette VR, it has error (error 60 0313) during exposure with large focal spot. Exposure for small focal spot is OK. Have you got any idea what is the cause of that?
In my opinion it is kv board or rotor board.
to rate and post a comment
Manuel Casanova
Re: X-ray jedi generator problem
February 13, 2007 03:51
Marcin:
I think that the problem is fixed in HV Tank or KV control board, the description error in service manual for silhouette, indicated as follows:
Invertar max, ILR current detected
In your case if small filament is ok, the rotor board is ok, I{m sure if you you change the KV control board yuopr system running.
Maybe check the data cable bus.
Best regards
to rate and post a comment
Staate Hayward
Re: X-ray jedi generator problem
February 13, 2007 05:57
I just checked the Proteus service manual (which is a Jedi generator as well) and agree that the problem is the KV control Bd. and worst case it could be the HV tank.
to rate and post a comment
necdet sahin
re: Re: X-ray jedi generator problem
April 20, 2015 10:07
hello
good work
Contact D x -ray machine v silhouette technical service guide is required.
What could be the problem in our popcorn IGBT device. Mail necexray@hotmail.co I
Thank you in advance
to rate and post a comment
Tayfun Canli
Re: X-ray jedi generator problem
February 18, 2007 05:55
Hi,
I really don't know GE Silhouette VR, but Jedi used by CT systems also as you know and some cases, Err 60-313h ILR Max current detected may caused by tube. If you have different error with old tube be sure your new tube is ok before order KV board. ( one of the filament may ok, but it will fail after a while if the tube is defective). Please try no load test if that kind of test is available at your system.
to rate and post a comment
yuvaraj sankaran
Re: X-ray jedi generator problem
February 19, 2007 08:11
Hi marcin,
For what reasons or errors you have changed the tube.
to rate and post a comment
Marcin Gorny
Re: X-ray jedi generator problem
February 21, 2007 02:51
Old unit didn't rotare.
Bearing had been broken.
When I conected high voltage cables to the old tube (in order to check errors) Erros was the same but inversly. It mean error is for small focal spot. Large is OK.
to rate and post a comment
Deon Armes
Re: X-ray jedi generator problem
March 01, 2007 05:16
Marcin,
This generator has diagnostic software you can us to test things on it. You need the Jedi software and a laptop cabable of running DOS and a serial port and cable. I havent had this problem but I know this unit will be difficult to troubleshoot and calibrate without the software. Good luck!
Deon
to rate and post a comment
Zahid Rustam
re: Re: X-ray jedi generator problem
February 25, 2013 10:07
hi i am looking for software for Jedi X-Ray Generator.shall be grateful if some one can help
to rate and post a comment
salman khan
re: re: Re: X-ray jedi generator problem
February 25, 2013 01:34
where is the system located?
to rate and post a comment
Zahid Rustam
re: re: re: Re: X-ray jedi generator problem
February 25, 2013 04:24
ABBOTTABAD PAKISTAN . IF YOU ARE FROM PAKISTAN , CAN U HELP REPAIRING GE VR SILHOUETTE RAD ROOM
to rate and post a comment
salman khan
re: re: re: re: Re: X-ray jedi generator problem
February 26, 2013 08:21
give me your contact number via email salmankhan.ge.ct@gmail.com
to rate and post a comment
Andrés Alzamora
re: re: re: Re: X-ray jedi generator problem
July 23, 2015 02:01
dear help me i have a xray system GE XR6000 and there error C03-0001 JEDI Communication Error, the cables is ok, the TAV board is OK, the console is oK too, but the problema follow
please write to aalzamoram@hotmail.com
Thanks
to rate and post a comment
Andrés Alzamora
re: re: re: Re: X-ray jedi generator problem
July 23, 2015 02:01
dear help me i have a xray system GE XR6000 and there error C03-0001 JEDI Communication Error, the cables is ok, the TAV board is OK, the console is oK too, but the problema follow
please write to aalzamoram@hotmail.com
Thanks
to rate and post a comment
tahar nouri
re: Re: X-ray jedi generator problem
February 26, 2013 08:23
Hi
I am lookig for this diagnostic software , if you have
pls send me
thanks in advance
to rate and post a comment
Youbissi Pascal
re: re: Re: X-ray jedi generator problem
February 27, 2013 08:41
Hi
I am looking for this diagnostic software,
pls send me
thanks in advance
to rate and post a comment
IrinaSaenz
re: Re: X-ray jedi generator problem
August 19, 2014 05:41
Hi:
I have a GE XR6000 xray system, and the console shows the C005-0001 error most of the time. I just can take head shots, but when i Change to lungs or chest or spine, this error appears and the anode doesnt even rotate. Sometimes when I change to stand bucky to table bucky allows the shot, but then fails. Can you help me please. I dont have the jedi software nor db9 cable configuration to connect and diagnost well. Please help me with this. My mail is: irina.saenz81@gmail.com
to rate and post a comment
Paulo Dick
re: re: Re: X-ray jedi generator problem
January 26, 2017 10:01
Hello! I need the Jedi Software!
Can anyone send to me? A quick software install tip would be awesome!
Thanks!
paulo@cpservice.com.br
to rate and post a comment
VERNE CHAPPELL
re: re: Re: X-ray jedi generator problem
June 21, 2017 04:21
Hi,I am having the exact same problem. Please let me know what your fix was.thank you.
to rate and post a comment
Maurice Dilick
re: re: re: Re: X-ray jedi generator problem
July 11, 2017 05:03
Which software is required? Mostly, there is software to upgrade to Jedi 5.47, but normally, one saves the previous data, then downloads onto non-volatile memory when changing the KV Control.
The manual is incorrect, can send a proper copy to your EMAIL
to rate and post a comment
antonio yallonardo
re: X-ray jedi generator problem
April 08, 2015 04:15
Please i have a trouble with the silhouete vr when try to get a shot nothin happen and no error from the sistem
to rate and post a comment
nour tal
re: X-ray jedi generator problem
April 26, 2015 11:57
X-RAY MVP 60/80/100 PROBLEM.
Hello.
I have an X-ray generator brand Ge 60/80/100 series. The console displays "Ext lock.". Would you help me understand this message? Thank you very much.
to rate and post a comment