i am getting XC error
code (0xb2640000)
dev (0x00010002)
all error
3b3f3f3f3f3f3f3f3f3f2f2f2f2f2f00

i am able to do IF adj upto 30ma beyond that this error comes again.any advise??

Log inor Register

to rate and post a comment

(53)
(11)
(16)

Dominique Bultot

re: XC error asteion multi

August 14, 2012 02:34

probably arcing, send me a mail to : dominique.bultot@skynet.be I will forward a soft to read Spelmann error on Asteion

Best rgs

PS : first of ALL , check HV plug : if dark and smell smoke, it's an arcing probably on HV cables

Log inor Register

to rate and post a comment

(69)
(6)

puma

re: re: XC error asteion multi

February 12, 2015 05:30

Hi Dominique,

can you forward me the sw to read spellman errors?

I have Asteion 4 slice system and have the same problem Error in XC.
code (0xb2640000)
dev (0x00010002)
all error
3b3f3f3f3f3f3f3f3f3f2f2f2f2f2f00
My email is tiralb123@gmail.com

Thanks

Log inor Register

to rate and post a comment

(9)

rodelo

re: re: re: XC error asteion multi

September 11, 2015 07:03

we have same problem, pls share if you have solved this

Log inor Register

to rate and post a comment

(9)

rodelo

re: re: XC error asteion multi

September 11, 2015 07:02

I have xc error during warm-up. code 0xb264000 dev 0x00010002

detail error 3f3f3f3b3f3f3f3f3f2f2f2f2f00

If we try to do a scan the gantry rotates but same error is indicated

Log inor Register

to rate and post a comment

(82)
(10)

Vu Toan

re: re: re: XC error asteion multi

September 16, 2015 08:01

Pls check connectors from AC-HV-Tube
After that check tube ....
vutoan_medical06@yahoo.com

Log inor Register

to rate and post a comment

(82)
(10)

Vu Toan

re: re: XC error asteion multi

September 16, 2015 08:01

You done IF for large focus and small focus ?
the system can warmup or not ?
The system can scan 1mm,3mm, 5mm or 10mm ?

With error 3b 0011xxxxb (Abnormal input power supply (AC Fault)
Pls check follow chart
AC fault
DOES FAULT
OCCUR DURING
EXPOSURE?
MONITOR TP13
ON THE SYSTEM
CONTROL BD
AC FAULT
DOES RAIL
MONITOR MEET
SPEC?
REPLACE
SYSTEM
CONTROL BD
END
ARE RAIL VOLTS
PRESENT?
MONITOR E6 AND
E7 ON THE AC
PWB
IS FAULT
PRESENT?
RESET
GENERATOR
NO
YES NO
REPLACE AC
PWB
YES
NO
REPLACE AC
CHASSIS
YES
YES
NO
I think you check all connectors from AC control to HV and tube. vutoan_medical06@yahoo.com

Log inor Register

to rate and post a comment

Sanchez Vicario

re: XC error asteion multi

August 16, 2012 09:38

I think the power of OPCONT is faulty.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Philips Achieva 1.5T MRI Spare Parts Available Looking for Philips Achieva 1.5T MRI Parts?

We offer reliable & cost-effective spare parts for the Philips Achieva 1.5T MRI system helping you minimize downtime & keep operations smooth. Wide inventory available. Quick turnaround time. Expert support. Worldwide delivery. Contact Limor@elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)