re: re: XC Error code (0xb2640000) dev (0x00010002)
March 25, 2022 10:02
is your ac fault from you using SPAG or SRU XC program? next look at your x-ray generator if the tube is at top(12 oclock) i there a unit at 3 oclock or weights. next issue your voltage on slip rings should is 200vac +/- 5% if 250vac there is voltage problem and may crowbar system. does the PDU have a standby/bypass lever ?
go to gantry control panel and tilt unit, if it goes wrong direction your unit is out of phase. This will also give possible error. check voltage on input power is it 200v ac
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
MTC
XC Error code (0xb2640000) dev (0x00010002)
March 07, 2022 08:50
We de-install Aq64 and reinstall almost 2 months later.
Now the problem Error in XC.
code (0xb2640000)
dev (0x00010002)
3b3f3f3f3f3b3f3f3f3f2f2f2f2f2f00
Please advice us,
BR
Dash
to rate and post a comment
Konstantin Kuznetsov
re: XC Error code (0xb2640000) dev (0x00010002)
March 08, 2022 10:20
Hi
Abnormal input power supply (AC Fault)
to rate and post a comment
MTC
re: re: XC Error code (0xb2640000) dev (0x00010002)
March 15, 2022 11:36
I check AC unit in another Aq64. The AC unit work normally.
Its output 254VDC
to rate and post a comment
ToshibaCT
re: re: XC Error code (0xb2640000) dev (0x00010002)
March 25, 2022 10:02
is your ac fault from you using SPAG or SRU XC program? next look at your x-ray generator if the tube is at top(12 oclock) i there a unit at 3 oclock or weights. next issue your voltage on slip rings should is 200vac +/- 5% if 250vac there is voltage problem and may crowbar system. does the PDU have a standby/bypass lever ?
to rate and post a comment
ToshibaCT
re: XC Error code (0xb2640000) dev (0x00010002)
March 10, 2022 09:29
go to gantry control panel and tilt unit, if it goes wrong direction your unit is out of phase. This will also give possible error. check voltage on input power is it 200v ac
to rate and post a comment
MTC
re: re: XC Error code (0xb2640000) dev (0x00010002)
March 15, 2022 11:36
I check the tilt, it was correct direction.
Any another idea ?
to rate and post a comment