XC error in Aquilion16

Leon Dang

XC error in Aquilion16

June 04, 2012 09:24

Hi everybody!

My system had problem after we changed one new heat exchanger for Aquilion16. When heat of X-ray tube become 0%, we warmed up and it displayed "error in XC (XC Nack) ; code (0Xb2500040) - dev (0x00010002) - pro(1)". Then we warmed up again and it could work on whole day without error. But few day later, error happened. Can anybody give me some advice?

Log inor Register

to rate and post a comment

(53)
(11)
(16)

Dominique Bultot

re: XC error in Aquilion16

June 04, 2012 11:33

I hope for you you be carefull with hydraulic plugs...maybe a bubble inside. But before to be afraid, connect the laptop to the XC test cable and tell me what you can check

Log inor Register

to rate and post a comment

Leon Dang

re: re: XC error in Aquilion16

June 05, 2012 09:26

Dear Dominique

XC log error almost is F7FFEFFFFF and sometimes is F7FFEF7DFF.

Thank you!

Log inor Register

to rate and post a comment

(44)

mcxrayca

re: re: XC error in Aquilion16

March 14, 2017 10:24

Hello Dominique, do you know the pinout of the cable to test the Xc?
I have aquilion 64 with a error code[0xb2640000] dev[0x00010002]
xc detil error 333f3f3a3f3f3f3f3f3f3f3f372f2f00

Log inor Register

to rate and post a comment

(2)

saood khan

re: XC error in Aquilion16

June 04, 2012 11:44

Dear send me xc error log.

Log inor Register

to rate and post a comment

(2)

saood khan

re: re: XC error in Aquilion16

June 07, 2012 11:33

Dear
Your heat exchanger is defective.

Log inor Register

to rate and post a comment

Leon Dang

re: re: re: XC error in Aquilion16

June 07, 2012 11:49

Thank you for your advice! It's useful for my case.

Log inor Register

to rate and post a comment

(53)
(11)
(16)

Dominique Bultot

re: XC error in Aquilion16

June 06, 2012 09:09

Error is "tube flow switch"...confirm my first idea : probably bubble inside the closed oil circuit when you connected the new heat exchanger. Or heat exch defective.

Log inor Register

to rate and post a comment

Leon Dang

re: re: XC error in Aquilion16

June 07, 2012 11:49

Yes, I checked and confirmed thermal switch in heat exchanger had problem. Maybe we need to claim new one. Thank you for your help!

Log inor Register

to rate and post a comment

(2)

David Meireles

re: re: re: XC error in Aquilion16

November 05, 2013 08:13

Hy i have the same error than LEON, but in detail XC i have F7FFFFFFFFFFFF. And if i go to the error log of the XC Maintenance, this error doenst mean anything?? This situation happens after one day after the full instalation of a new tube!
I check cable, HV Cable , Block test and all seems be ok.

Any idea??

Tks for help

Log inor Register

to rate and post a comment

(2)

David Meireles

re: re: re: XC error in Aquilion16

November 05, 2013 08:13

Hy i have the same error than LEON, but in detail XC i have F7FFFFFFFFFFFF. And if i go to the error log of the XC Maintenance, this error doenst mean anything?? This situation happens after one day after the full instalation of a new tube!
I check cable, HV Cable , Block test and all seems be ok.

Any idea??

Tks for help

Log inor Register

to rate and post a comment

(88)
(36)

Anton Spaci

re: re: re: re: XC error in Aquilion16

November 07, 2013 03:15

means HFG error , what was the original error make you change the tube ?
Did you change the HV cable ?

Log inor Register

to rate and post a comment

(6)
(1)

EngDM

re: re: re: re: re: XC error in Aquilion16

November 09, 2013 07:42

Tks for the repply Anton. I changed the tube because i was under KV and sometimes under mA, and impossible to make one scan.
No i didnt changed the HV Cable. I putted more special oil in the HV Cable (tube side), i checked all the HV connections in inverters and add more varian silicon. They worked with the system now, but they have one or two errors per day. They made XC Reset and can follow normally the exams. We also changed the XC Board.

Do you think that the HV Cable (SRU-H to tube) are the issue???

Tks

Log inor Register

to rate and post a comment

(88)
(36)

Anton Spaci

re: re: re: re: re: re: XC error in Aquilion16

November 13, 2013 06:51

Sometimes it take times to stabilize the tube , it is very important to make seasoning while you replace the tube and maybe you have to do it again and readjust if

Log inor Register

to rate and post a comment

(88)
(36)

Anton Spaci

re: re: re: re: XC error in Aquilion16

November 07, 2013 03:15

other trick , when you get error with xc test , in staus info you can see with staus N°2 if you have a broken inverter

Log inor Register

to rate and post a comment

(6)
(1)

EngDM

re: re: re: re: re: XC error in Aquilion16

November 09, 2013 07:42

Ah regarding this, i think normally you can just pay attention to the status nº2 when you have the ST status with 01, right???

When we got this XC error we dont have special error and :
ST are 00,
Status 2 : 04000

i think this means that we dont have problem with inverters??? Meanwhile i made all the tests with block test and with no erros??

More one time do you think that the issue could be the HV cable!!

Tks

Log inor Register

to rate and post a comment

(88)
(36)

Anton Spaci

re: re: re: re: re: re: XC error in Aquilion16

November 13, 2013 06:51

Hi we don t care about status 1 but i think but not sure that normal status for the second should be 7FFF
If i was at your place i will change my HV cable
good luck

Log inor Register

to rate and post a comment

(6)
(1)

EngDM

re: re: re: re: re: re: XC error in Aquilion16

November 21, 2013 08:20

Anton, tks for the repply.. Recently i didnt have time to test a new HV cable because the system had other problem with the servo amplifier of the DD motor. And just after that i will test with a new HV cable.. I will let you know about this...tks a lot

Log inor Register

to rate and post a comment

(44)

mcxrayca

re: re: re: re: re: XC error in Aquilion16

March 14, 2017 10:25

Hello Anton, please you know the pinout of the cable to connect to a test the Xc

Log inor Register

to rate and post a comment

stats