I am working in a XR6000 the equipment is not generating x ray, the system turn on, but no emision of x ray, the anode is not rotating, and only appears the error 020-0253, and sometimes the error 020-0212, some suggestion?? my experience in this equipment is poor.
Thanks in advanced.
code 50
50-0212H
MAINS_DROP detected.
The firmware of the Heater board has detected the mains_drop signal activation and has transmitted error to kV control
1.Interference (spikes)
2.Mains drop
3.Cable or connector contact in DC bus between AC/DC and auxiliary
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
ENRIQUE RIOS
XR6000 error 020-0253
June 05, 2020 09:41
Hi Greetings.
I am working in a XR6000 the equipment is not generating x ray, the system turn on, but no emision of x ray, the anode is not rotating, and only appears the error 020-0253, and sometimes the error 020-0212, some suggestion?? my experience in this equipment is poor.
Thanks in advanced.
to rate and post a comment
juan pablo amado
re: XR6000 error 020-0253
June 08, 2020 11:10
Code: 20
Heater Warning
0253H
No CAN message received within 4 secs
code 50
50-0212H
MAINS_DROP detected.
The firmware of the Heater board has detected the mains_drop signal activation and has transmitted error to kV control
1.Interference (spikes)
2.Mains drop
3.Cable or connector contact in DC bus between AC/DC and auxiliary
to rate and post a comment
KOHOUNKO ULBRITCH
re: XR6000 error 020-0253
June 10, 2020 11:16
check the anode board
to rate and post a comment