Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

Xray testing light field alignment

(1)

Ed Harris

Xray testing light field alignment

April 09, 2018 09:38

Does anyone know if using an 8 x 8 field of view size would result in a better light field alignment test than using a 12 x 12? Does the size of the field of view matter?

Log inor Register

to rate and post a comment

(97)
(14)

Donald Bradbury

re: Xray testing light field alignment

April 10, 2018 08:52

Not particularly. I always used a 4X4 field size with Pretty good results. I wouldn't recommend going much smaller than that. The important part is the coincidence of 2% or better of the SID.

Log inor Register

to rate and post a comment

(28)
(1)

Snoclimber

re: Xray testing light field alignment

April 10, 2018 08:52

No. what will be more critical is your SID. each manufacture has their own spec. Some have a +/-x mm at a set sid. Others may have a certain percentage of FOV at a set SID.

A Universal approach would be to use a wall bucky at 40cm and adjust to a knats arse or better than 1/2 of manufactures spec. I am now using the raysafe light field ruler which really speeds up the process

Log inor Register

to rate and post a comment

(28)
(1)

Snoclimber

re: Xray testing light field alignment

April 10, 2018 08:52

usually 2% of SID at 100cm but you should be able to do way better than that

Log inor Register

to rate and post a comment

(77)
(1)
(4)

eduardo medina

re: Xray testing light field alignment

April 10, 2018 08:52

It dosent matter the size

Log inor Register

to rate and post a comment

(62)
(14)

MM

re: Xray testing light field alignment

April 10, 2018 08:54

I use normaly 20x20cm with good results. Then I use a 24x30cm cassette to check X-Ray
MM

Log inor Register

to rate and post a comment

stats
DOTmed text ad

JOHNSON & JOHNSON MONARCH QUEST - For Sale $299k or Lease $6k Per Month, 48 Months

Lightly used system available direct from the end user customer. MONARCH QUEST Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.

stats

Upcoming Auction Events

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)