I'm having a problem with a GE Hispeed FX/i system. I made a LFC with no errors, when try to do Q-Cal calibration I got an error,
30-1010-01 : CIF Zero Das Data
Can not finish Q-cal, the system gives only the first 6 exposures, when try to make the warm up, it gets the error.
I checked:
- Das power supply, where exchanged a defective -12V PS.
- CIF board replaced
- Das data transfer test, ok
- Cam boards swapped
- Cam boards test by turning on the switch 3-8 of cif board, giving a noisy curve but no errors on any channel. I have a picture of that curve
- There is image on the screen, but a lot of artifacts,
At this point I beg some help, some light in the darknessType your question or comment here...
Hi,
I hope you can help me with a GE CT Brightspeed 16. I am new with GE CT.
channel 526 (detector 590) Raw 5B=Converter 66 is reported as broken when try to do Fastcal.
Using "Scan Analysis" a very big peak is shown on graphics.
After exchange converter boards same channel shows such peak.
There is not possible to do DAS Gain Calibration because this bad channel.
I an trying to get a Converter crossover cable to do more test. It will take some time.
On Siemens CT I know that there is a way to "kill/hide" channels that are bad. This can be done manual and/or automatically (Defective channel function). Do you know if there is any tool/way to do the same with GE CT?
Lightly used system available direct from the end user customer. MONARCH QUEST
Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
Fabian Guerrero
Zero Das data on Hispeed FXi CT
May 20, 2016 07:47
I'm having a problem with a GE Hispeed FX/i system. I made a LFC with no errors, when try to do Q-Cal calibration I got an error,
30-1010-01 : CIF Zero Das Data
Can not finish Q-cal, the system gives only the first 6 exposures, when try to make the warm up, it gets the error.
I checked:
- Das power supply, where exchanged a defective -12V PS.
- CIF board replaced
- Das data transfer test, ok
- Cam boards swapped
- Cam boards test by turning on the switch 3-8 of cif board, giving a noisy curve but no errors on any channel. I have a picture of that curve
- There is image on the screen, but a lot of artifacts,
At this point I beg some help, some light in the darknessType your question or comment here...
to rate and post a comment
Borhan Kalash
re: Zero Das data on Hispeed FXi CT
May 21, 2016 04:02
There is a dip sw for disabling the zero das on the CIF as i remember (or the board near it)
You can find its location in the manual
to rate and post a comment
Fabian Guerrero
re: re: Zero Das data on Hispeed FXi CT
June 04, 2016 03:39
thanks for your help friend
to rate and post a comment
Ajeesh Nair
re: re: re: Zero Das data on Hispeed FXi CT
June 14, 2016 10:16
Hi Fabian . .. Is your issue solved ...
to rate and post a comment
Fabian Guerrero
re: re: re: re: Zero Das data on Hispeed FXi CT
June 29, 2016 10:18
Hi Ajeesh, exchange position of the cam board and it works. but sometimes noisy channel appears on the screen
to rate and post a comment
Ajeesh Nair
re: re: re: re: re: Zero Das data on Hispeed FXi CT
June 30, 2016 09:37
Hi Fabian, Seems there is a faulty cam board. If you have some phantom image with artifacts please mail to
ajeesh.nair@medirays.in
to rate and post a comment
Borhan Kalash
re: re: re: Zero Das data on Hispeed FXi CT
June 29, 2016 12:46
did you solve the trouble
to rate and post a comment
Mr.CT.service
re: re: Zero Das data on Hispeed FXi CT
March 09, 2018 09:29
Hi,
I hope you can help me with a GE CT Brightspeed 16. I am new with GE CT.
channel 526 (detector 590) Raw 5B=Converter 66 is reported as broken when try to do Fastcal.
Using "Scan Analysis" a very big peak is shown on graphics.
After exchange converter boards same channel shows such peak.
There is not possible to do DAS Gain Calibration because this bad channel.
I an trying to get a Converter crossover cable to do more test. It will take some time.
On Siemens CT I know that there is a way to "kill/hide" channels that are bad. This can be done manual and/or automatically (Defective channel function). Do you know if there is any tool/way to do the same with GE CT?
Best regards.