Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Ziehm exposcop 7000 error E65

(24)
(1)

Marie Martinez

Ziehm exposcop 7000 error E65

November 02, 2011 09:09

I got an error E65 and the service manual says to call customer service, has anyone seen this problem please help. Contact me at pmeconsulting@verizon.net

Log inor Register

to rate and post a comment

(37)
(6)

John LoPresti

Re: Ziehm exposcop 7000 error E65

November 18, 2011 08:37

That error 65 means there is no video signal arriving at the Module 3. The following are possible causes....
Cable interrupted
Check video cable from CCD camera to Module 3
Check video connectors of camera and Module 3
I.I. power supply is set to 6 V camera supply voltage, camera requires 12V
Set camera voltage to correct value (J1 on U316)
see Module4 → Ch. , p. 7-8
Camera has no supply power
Check camera supply voltage
If 18 V input voltage of the i.i. power supply is present and no camera voltage
is present, replace i.i. power supply
CCD camera defective
Replace CCD camera
Module 3 input stage / genlock defective
Replace Module 3

Log inor Register

to rate and post a comment

(24)
(1)

Marie Martinez

Re: Ziehm exposcop 7000 error E65

November 24, 2011 11:02

Thanks it was a couple of bad video cables from the II to the camara.

Log inor Register

to rate and post a comment

(37)
(6)

John LoPresti

Re: Ziehm exposcop 7000 error E65

November 25, 2011 11:54

glad to be of help

Log inor Register

to rate and post a comment

(37)
(6)

John LoPresti

Re: Ziehm exposcop 7000 error E65

November 25, 2011 11:54

your welcome

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)