ultrasoundman
Zonare Z.One
October 22, 2018 02:47
I have a client with a Zonare Z.One it is displaying as he said triangles or maybe missing vectors. Only small portion of the screen shows the 2D Image. Any suggestions and or Service manual would be helpful
to rate and post a comment
Stanley Jackson
re: Zonare Z.One
October 23, 2018 09:06
Is it the display from the probe image or the entire screen that's showing triangles? If it's the probe image then the probe is bad, (i.e. missing elements). If the entire screen is being affected then you need to replace the Scan Engine. Contact Zonare, they are pretty good with tech support and will clear this up for you.I do not have a manual but worked with them on multiple occasions.