The code states: No acknowledgement message from Slave CPU for 3 communication trials. Checks: Turn pump off then back on. Corrective action: if failure recurs replace CPU board.
Ophthalmic-focused expertise in YAG systems including repair, maintenance and calibration handled with precision by trained professionals. Request service • sales@laserlocators.com • 1-877-924-2020
Teri Rufner
Baxter 6201 infusion pump error code F_82
August 16, 2016 09:23
What is the F_82 error code? I haven't seen it before.
to rate and post a comment
John Holt
re: Baxter 6201 infusion pump error code F_82
August 17, 2016 10:40
Error Code F_83 means The CPU needs to be replaced if it cannot be reset by turning the unit off and back on.
Good Luck....
to rate and post a comment
Donghoon Koh
re: Baxter 6201 infusion pump error code F_82
August 17, 2016 10:44
"No acknowledgment message from Slave CPU for three communication trials." in Service Manual.
checks : Turn device off and back on.
Corrective action ; If failure recurs, Replace CPU board.
to rate and post a comment
William Clarke
re: Baxter 6201 infusion pump error code F_82
August 17, 2016 10:48
Bad CPU board
to rate and post a comment
Bill Tripp
re: Baxter 6201 infusion pump error code F_82
August 17, 2016 10:48
The code states: No acknowledgement message from Slave CPU for 3 communication trials. Checks: Turn pump off then back on. Corrective action: if failure recurs replace CPU board.
to rate and post a comment