Error at US hd11xe philips

(24)
(4)

Linda Tirado

Error at US hd11xe philips

February 20, 2012 10:22

We get an applcation error when booting system, the message appearing when error comes out says that initialization from report creation packge has failed and this cause problems at the patient database functions.
We restarted the system letting some intervals of even 30 minutes having the same results.
When selecting patients we get another message:
The report packages is starting,this action will end in 30 sec.
If you have transfered the system on another netwrok this could take 2 minutes.
Patient, review, report, acquire will not be available until initialization ends.
Try again.

And this error code appears:
Error ID: MercCert:Guistartimeout

Please, any suggestion on the solving this, the messages appears in spanish so I translated content.
Thanks!

Log inor Register

to rate and post a comment

Jim Carr

Re: Error at US hd11xe philips

February 21, 2012 12:06

It sounds like you may have a corrupted Preset or Study. Try restoring the System and user presets. Then try deleting all the studies. Patient study files may be corrupted but not show up in the study list. There may be a utility to reset the Study database, similar to the "HD Reset" on an iU22, although there isn't a reference to such a command in the service or user manuals I have.

Jim Carr
Unisyn

Log inor Register

to rate and post a comment

(24)
(4)

Linda Tirado

Re: Error at US hd11xe philips

February 23, 2012 10:26

Thank you very much to you all!

Log inor Register

to rate and post a comment

(11)
(3)

Phillip Wood

Re: Error at US hd11xe philips

February 21, 2012 12:07

Hard drive and reload of software should fix it.

Log inor Register

to rate and post a comment

(36)
(14)

Omar Ortiz

Re: Error at US hd11xe philips

February 21, 2012 12:07

it looks for me you have a corrupt software. This units soffer of software corruption a lot. with a simple software reload you ahould solve this issues. you need to have a good backup and you also need to have the IP adress from the computer because this is not at the backup, after the reload you will need to set it manually.

hope this helps,

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

Re: Error at US hd11xe philips

February 21, 2012 12:08

I believe you will have to reload the software, please contact me ricardo@ultrasolutions.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Don’t Compromise on Safety or Speed

From skilled GE depot repairs to patient-ready accessories, Southeastern Biomedical specializes in GE to keep your facility safe, efficient, and cost-effective. 888-310-7322.

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)