Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Lorad M IV consol shuts down when exposing

(3)

Lars Nielsen

Lorad M IV consol shuts down when exposing

February 23, 2012 09:40

I have a customer that uses a Lorad M IV, when performing a exam, the consol shuts off when trying to expose, then a restart is necassery.

Have you got any idea on what could be wrong?

lcn@dumed.dk

Log inor Register

to rate and post a comment

(36)
(14)

Omar Ortiz

Re: Lorad M IV consol shuts down when exposing

February 23, 2012 05:09

hi, this happen once the exposure start, or happen when the tube rotor start but no x ray have start yet? Any error was generated when the problem occurs? from what year is the unit? from what year is the x ray tube? this looks for me like arquing of the high tension (x ray tube or high tention generation) this is a new install or the unit was working ? if it is a new install this could be a calibration problem or if the unit was under use this looks like arcquing. the incoming power didn't change? if you incoming power is not right this could be arquing problem. hope this helps,

Log inor Register

to rate and post a comment

(50)
(4)

Douglas Gray

Re: Lorad M IV consol shuts down when exposing

February 24, 2012 12:06

Does the unit come back on when you power up ? I had a problem with the rotation switch sometimes it would not come back on sometimes it would.

Log inor Register

to rate and post a comment

(44)
(3)
(1)

pablo neira

Re: Lorad M IV consol shuts down when exposing

February 26, 2012 08:27

sr you need service manual i can scan you pdf format send you usd 300
skype paulnelsonneiragarrido

Log inor Register

to rate and post a comment

(54)
(7)
(9)

Deborah Breland

Re: Lorad M IV consol shuts down when exposing

February 27, 2012 04:56

Check the 5 volt coming from the low voltage power supply on the console and make sure it is between 5 to 5.1 VDC. If it not adjustment is needed or Low Voltage Power Supply is bad.

Log inor Register

to rate and post a comment

(44)
(3)
(1)

pablo neira

Re: Lorad M IV consol shuts down when exposing

February 28, 2012 01:02

dear sir i have this manual service in paper i can scan you please contact to me mantroequip@yahoo.es

Log inor Register

to rate and post a comment

(11)

Cale Clinkenbeard

Re: Lorad M IV consol shuts down when exposing

April 13, 2012 07:14

Check the solid state relay, suppressor on it, and the contactor. All located in the lower left of the gantry unit.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)